how do you go about doing a ZS test with a fluke 1652c
I know how to get the RE reading but it's coming out a bit strange, a bit of advice would be great.
 
Did you invent your username from what your fluke said on the screen Err4....

Just turn the knob ttote trip or no t setting, no trip if inline with an rcd.

lug your leads in or socket adaptor, then press the big button...
 
If you are measuring Re this is for earth rods on a TT.
To measure Zs use F3 to remove Re.

From the manual:

No trip, RCD in circuit
1. Turn the rotary switch to the Z| NO TRIP position.
2. Connect all three leads to the L, PE, and N (red, green, and blue)
terminals of the tester.
3. Press F1 to select L-PE. The display shows the ZL and no trip
indicator. (Do not press F3, as this will select RE)
4. Before you do a loop impedance test, use the zero adapter to zero the
test leads or the mains cord. Press and hold ‘zero’ for more than two
seconds until the ZERO annunciator appears. The tester measures
the lead resistance, stores the reading in memory, and subtracts it
from readings.
5. Connect all three leads to the L, PE, and N of the system under test or
plug the mains cord into the socket under test.
6. Press and release ‘Test’. Wait for the test to complete.
The primary (upper) display shows the loop impedance.
7. To read the Prospective Earth Fault Current, press the F3 key and
select lK. The Prospective Earth Fault Current appears in amps or kilo
amps in the secondary (lower) display.
8. If the mains is too noisy, Err 5 will be displayed. (The measured value
accuracy is degraded by the noise.) Press the down arrow to display
the measured value. Press the up arrow to return to the Err 5
display.
This test will take several seconds to complete. If the mains is
disconnected while the test is active, the test automatically terminates.


Hi current
If no RCDs are present in the system under test, you can use the high current
Line Earth (L-PE) loop impedance test.
1. Turn the rotary switch to the Z| TRIP position.
2. Connect all three leads to the L, PE, and N (red, green, and blue) terminals
of the tester.
3. Press F1 to select L-PE. The trip symbol appears to indicate that hi current trip
mode is selected.
4. Press F2 to select between Ω and mΩ resolution for the test results. The
mΩ resolution test takes between 30 and 60 seconds to complete.
5. Connect all three leads to the L, PE, and N of the system under test or
plug the mains cord into the socket under test.
6. Press and release ‘Test’. Wait for the test to complete.
The primary (upper) display shows the loop impedance.
7. To read the Prospective Earth Fault Current, press the F3 key and
select lK. The Prospective Earth Fault Current appears in amps or kilo
amps in the secondary (lower) display.
8. If the mains is too noisy, Err 5 will be displayed. (The measured value
accuracy is degraded by the noise.) Press the down arrow to display
the measured value. Press the up arrow to return to the Err 5
display.
 
Sorry guys I have a 1652 not version c so I only have 6 choices and only one is z1 loop I can't understand how to do the zs test, the reply above is not for my meter.
 

Reply to Fluke 1652c ZS test in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

Similar Threads

Hi All Happy new year to all! First post but long timer lurker, so thanks for all the previous help! Just wanted to clarify something I have...
Replies
7
Views
810
Hi All New to this forum, have read the posts on here from google but only recently signed up. I'm having some issues and some input would be...
Replies
13
Views
1K
Now Ive done Periodic Inspection and Initial Virification Qualifications with LCL/EAL (so non C&G 2394/95) # strange part building Modules , this...
Replies
6
Views
935
I got a TEXT message (i dont know them and i cant find out who they are from the phone number) "Hi speaking from Indeed Are you available to give...
Replies
15
Views
845
Hi all, Testing question for you. I'm currently testing an existing domestic installation. The wiring is about 20-25 years old. We have rewired...
Replies
7
Views
1K

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc
This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock