Taken from page 6 of the KT63 manual;

Special polarity test function.
It is a little known fact that a system can be reverse wired with Line (Phase) to earth/neutral and earth/neutral to Line (Phase).

The sockets will all work and conventional loop testers will show and test that everything is correct despite this very dangerous wiring condition.

Although extremely rare, this miswire condition can exist so if your test shows this fault do not proceed – if in any doubt advise your customer to contact their supply company immediately.

Audible tones
A simple selection of audible tones is used to supplement the visual display. These help the user by providing intuitive feedback during testing. In addition to warning about dangerous or unstable supply conditions they provide a very quick confirmation that the measurement process is taking place and, upon completion of the test, a warning if the results are likely to be regarded as a failure.

The meaning of the tone for each individual function is covered in detail in the relevant
section. Generally however there are five types of tone emitted.

Danger
A rising siren type alarm
In the event of a potentially dangerous situation such as connecting to a live supply when configured for insulation
testing. Will be accompanied by the Red Voltage/Polarity warning LED flashing.

Warning
A continuous 2 tone alarm
An unsuitable supply configuration such as a mains supply with incorrect polarity or having the leads connected wrongly will be accompanied by the Red Voltage/Polarity
warning LED flashing.

Wait-Test in progress
A steady beeping sound Emitted whilst a measurement is in progress. The same
tone is sounded when used in Handsfree mode to indicate that continual measurement is being made

Test completed
A single beep
Sounded upon completion of ameasurement to indicate that the result is being displayed

Alert
A short 2 tone alarm
Sounded when a test returns a result that is likely to be regarded as a failure e.g. An insulation test that gives a result of less than 2 MΩ
 

Attachments

  • KT63_manual.pdf
    1 MB · Views: 2
I think now would be the time to do the proper polarity test to see what's what.
It’s a difficult job as it‘s remote and for an insulation company who just want the outlets off ASAP. I’ve never had a warning light come on with any other job. Will look into checking polarity and fixing the issue when we return. All feedback much appreciated! ?
 
It’s a difficult job as it‘s remote and for an insulation company who just want the outlets off ASAP. I’ve never had a warning light come on with any other job. Will look into checking polarity and fixing the issue when we return. All feedback much appreciated! ?

It doesn't take long to check polarity when you've lareday got the tester out and connected to the socket. Just set the mode to voltage measurement and change the test leads around a couple of times.
 
(Assuming you are testing sockets ,from mentioning plugging in)
Do you know if a ring / radial ?
(My Fluke will freak out , if "you" are picking up stray voltages - leaning against culprit)
 
Last edited:

Reply to Polarity voltage warning ⚠️ in the UK Electrical Forum area at ElectriciansForums.net

Similar Threads

Today I went to fix a minor loose screw on something and after switching off, etc, got out my Martindale VT12 continuity & voltage tester and...
Replies
3
Views
420
A BG fused connection unit, with neon. A boiler. Fuse out. Spur switched off. I was checking polarity after a colleague had 2nd fixed it and...
Replies
7
Views
710
Hello all. So I am sure the expertise on this forum will be able to decode this very easily. But it has completely baffled me. I'm in training so...
Replies
14
Views
575
Hi all, Firstly I am new to the forum, so will start by saying hi. I work as an electricians mate, and am joining the forum to gain experience...
Replies
3
Views
618
Hi all, New to the forum. I have been asked to look at this for one of our guys who's had an issue onsite after some electrical works had been...
Replies
4
Views
675

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Electrical Forum

Welcome to the Electrical Forum at ElectriciansForums.net. The friendliest electrical forum online. General electrical questions and answers can be found in the electrical forum.
This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock