Exemptions do exist but if someone is looking for a way around fitting an rcd then they are not designing their circuits correctly unless we are talking about feeding a socket for a kidney dialysis machine but other measures would also be used to avoid the socket been used for normal use.
Actually DW, a Haemodialysis machine requires a 30mA RCD in the supply.
It's in MI's and often the Health Authority spec, you would hope that Fresenius and Gambro are a bit more clued up than your average Far Eastern tat supplier.
I would have to check in MEIGaN and HTM06, mind this is slowly being replaced by Section 710.
A dialysis machine has an inbuilt battery backup, or at least the ones I work with do.
This will ensure that the patient will not loose a circuit, loosing a circuit now and again is not such, an issue, I was with a patient yesterday who lost a circuit on a Fresenius machine.
Not due to the machine though, and in minimal care, he had a blowout.
Even if the machine shuts down, the circuit can be manually pumped back into the patient anyway.
Dialysis whilst life saving, is NOT at such a critical level as life support.
If you look at the changes in section 710 then you will see the change in philosophy.
Whilst in the hospital environment you would probably find that the machine is on the essential supplies.
This is also partly to do with the issue that in the event of a power failure, the care staff can't be sorting out perhaps 40 patients at the same time, not that they are life support.
I am looking at one HHD install tomorrow morning, and I will be putting it on a 30mA RCD at the "origin" of "our" responsibility.
Whether the home has any existing RCD protection or not.
As far as sockets for normal use go, linked to HHD, then it is not an issue requiring anything different to normal really.
The patients are trained and instructed what and what not to do.
OK, Fresenius require EN 60309 socket outlets, but Gambro require BS1363.
In a hospital environment then you have ESS & non ESS so that covers that, however, often both are BS1363.