plugsandsparks

Electrician's Arms
Messages
1,626
Location
Chesterfield
My MFT failed calibration on some insulations resistance measurements but i am a little confused. I used a different more local calibrater as i needed the least downtime but his certificate is very different from the manufacturers one that was done before (now i have had time to compare side by side)

Basically my tester has failed 4 tests (one on 500V range and three 1000V range) of these 4, 3 readings are at a different much higher resistance tests than the manufacturer used before.

The only reading that is of comparison that failed is in the 1000V range and was marginal

So if i return it to the manufacturer i can see it simply being returned without needing anything doing.

Question: do i go back to calibrater and ask why has he tested at settings different from manufacturer (and harder to be accurate test modes)
Or do i send it back to manufacturer and just get it "done" ( i think i have been anyway)

Do i carry on regardless as the areas it has failed "1000V" range is not exactly used alot (at all really)
and the one on the 500V range is nothing that would make any difference to anything an electrican needs to measure on cable - however my tester has a new sticker saying calibration due august 2013 with fail written by hand in pink..... puzzled.

What do you think guys - always used manufacturer before and it just comes back OK.... this is new territory.
 
K

Knobhead

If you’ve sent it for calibration I would expect the test house to make any adjustments to bring it in to accepted range.

I used Dowding and Mills when I was working, (the company paid).
 

Archy Styrigg

Mentor
Electrician's Arms
Trainee Access
Messages
10,141
Location
Manchester
Did they even attempt to get it within range?
As Tony says, that's the whole purpose of calibration, to bring the instrument within limits! Not just test it!

If they've attempted to calibrate it and can't, it is probably in need of repair.
Get back to your so-called calibrators.
 
OP
P

plugsandsparks

Electrician's Arms
Messages
1,626
Location
Chesterfield
I was open mouthed when they told me, i did say "how can it fail calibration" and i did ask say to them " I thought you just made an internal adjustment" but they said no they dont do that, but did put in fresh batteries.
Going to ring Kewtech tommorow and see what they say.
Need the tester all week, then go on hols.
 

Archy Styrigg

Mentor
Electrician's Arms
Trainee Access
Messages
10,141
Location
Manchester
I was open mouthed when they told me, i did say "how can it fail calibration" and i did ask say to them " I thought you just made an internal adjustment" but they said no they dont do that, but did put in fresh batteries. Wuppee frigging do!
Going to ring Kewtech tommorow and see what they say. Good idea.
Need the tester all week, then go on hols.
Hope you get sorted.
Let us know what Kewtech say.
 
K

Knobhead

P&S, for other members of the boards information what's the name of the test house you used?
 
OP
P

plugsandsparks

Electrician's Arms
Messages
1,626
Location
Chesterfield
Not happy about stating test house at the moment as i am slightly embarrassed about my expectations that it would come back "sorted" and i need answer to e.g.

IR on 500V - applied values were 1M, 10M, 50M, 100M, 500M and 1000M, the one which it failed on was 1000M were it read 653M.

However last calibration done by Kewtech on 500V setting was 1M, 10M and 100M

quite frankly i dont give a monkeys about anything over 100M, think they are trying to be too clever.

Its the same story on 1000V setting, applied value 1M, 10M, 100M, 500M and 1000M
readings were 1.00M, 9.88M, 93.7M (fail), 452M (fail) and 620M (fail)

I have checked the specs of the tester and its claims of accuracy and this testing house is doing it literally by the book but them why does the manufacturer not do the same and why have a tester trying to accurately measure stuff like a 1000M
FFS

Anyway will talk to Kewtech in morning and report back
 

telectrix

Disrespected Scouser
Electrician's Arms
Trainee Access
Messages
63,488
Location
cheshire/staffordshire
strange, because most testers have a max of either 199meg or 299 meg. anything over is a >XXX
 
OP
P

plugsandsparks

Electrician's Arms
Messages
1,626
Location
Chesterfield
Mines KT65 auto ranging 20/200/2000 so its actually claiming this level of accuracy.
However even last job i did was still getting >1999 on 500V and a really long cable run almost got it down to 1700M
 

topquark

Electricians Arms Member
Mentor
Electrician's Arms
Trainee Access
Messages
3,377
I would be tempted to ask the "calibration company" what they would charge you to fix it. If they give you a price, then they're trying it on!! If it's capable of being fixed they should have already done it for what you've paid (unless they're own T's n C's say otherwise, but that would be unusual).
 

ruston

Respected Member
Mentor
Electrician's Arms
Trainee Access
Messages
5,537
Location
Northumberland
I would be tempted to ask the "calibration company" what they would charge you to fix it. If they give you a price, then they're trying it on!! If it's capable of being fixed they should have already done it for what you've paid (unless they're own T's n C's say otherwise, but that would be unusual).

Yes I agree with that , after all you have sent it for calibration not a check, you could have done that on your check box if that was the case.
 

thekingiam

Regular EF Member
Messages
1,488
Location
Uk
calibrate (k l -br t ) 1. To check, adjust, or standardize a measuring instrument, usually by comparing it with an accepted model.
 

thekingiam

Regular EF Member
Messages
1,488
Location
Uk
is there a certain standard for calibration ukas or something is it worth checking to see if they registered and maybe check ukas to see what standards are required for calibration
 
OP
P

plugsandsparks

Electrician's Arms
Messages
1,626
Location
Chesterfield
Good news, spoke to Kewtech this morning, they arranged for me to take it in, drove to Donny, had breakfast and returned to new calibration cert showing all good.
Tester was OK, you just needed to hold the button down for the readings to stablise on the higher IR ones, i do that as a matter of course but looks like the calibration firm recorded the first reading to come up rather than wait for it to stablise. The engineer gave me a demo of what was going on.
Have to say Kewtech were brilliant, second time in 2 years i have been there in distress and left both times fully fed and happy.
No charge was raised, the manufacturer and calibration company were in contact with each other and i have left them to resolve these types of issues.
So happy again
 

topquark

Electricians Arms Member
Mentor
Electrician's Arms
Trainee Access
Messages
3,377
Good to know that there's still good customer service to be had, credit to Kewtech!
 
OP
P

plugsandsparks

Electrician's Arms
Messages
1,626
Location
Chesterfield
Result!
Do a global PM with this so-called Cal House's name so we can avoid it! :)
No really, they are new to the game, the outfit have helped me out in the past i wouldn't want to even go there with friggin google watching every move. Its a transient issue which i am sure wont be repeated.

Not in the business of damaging reputations just because i have had an inconvenience and a gallon of diesel used up. I was more concerned about the efficacy of any work i do this week, but alls good.
If you dont change your calibration company from last year you wont meet them
 

KevinF

Active EF Member
Messages
37
Location
Aberdeen
Excellent result .
Kewtech customer service has regained them a customer .
As Tony said I hope you chase the other crowd for a refund
 