whats the improvements in the 1720 over the old 1552?
does it do non trip loop testing faster?
Additional features of the Megger MFT1720 compared to the previous model MFT1552:
Safety:- IEC61010-1 Category 4 (MFT1552 Category 3)
Insulation:- Adjustable buzzer threshold for preset insulation failure level.
Continuity/Low Resistance:- 15mA test.
RCD:- Programmable RCDs and 3 Phase RCD (No Earth) tests.
Loop:- 2 wire tests, Max Zs, R1+R2 and Touch Voltage values.
Phase:- Sequence test.
Other:- Switched probe included, Leakage current measurement (requires additional clamp).
[FONT=&]
Conclusions:
Some improvements and additions, but the ballyhoo of the press release was largely over-blow hyperbolae. The much trumpeted features, of dubious practical value, include the 15mA continuity/low resistance test, Max Zs and R1+R2. The 2 wire loop test is best described as
not fit for purpose at present, but each firmware update brings a slight improvement, this may yet prove to be useful when it’s sorted.
No-trip:
The No-trip loop measurement is slightly faster due increased processor speed, but not significantly so.[/FONT][FONT=&] But the good news is they still employ the patented multi-pulse, noise cancellation measurement method. It is because of the use of this technique that the Megger MFTs tend to be slower on No-trip, due to the addition processing required. Because of this they get unfair criticism from users of other MFTs (Fluke devotees in particular) for their apparently slow "No-trip" test, however, the Megger users get the last laugh because they very rarely have their test inhibited by a noise error, in fact it's so rare the majority of Megger users have never seen the noise error symbol on the display. Whereas Fluke users see their equivalent, Error 5 regularly, even in relatively noise free domestic environments[/FONT], thus requiring repeat testing.