I think Kewtech are the people who will know, why don't you give them a ring? 01522 546888 (Technical Support).
On their old KT45 dedicated loop tester, you could (from the KT45 manual):
If you wish to measure the Phase-Neutral loop impedance, then same procedure should be followed except the earth clip should be connected to the neutral of the system i.e. the same point as the black neutral probe.
If the system has no neutral then you must connect the black neutral probe to the earth i.e. same point as the green earth clip. This will only work if there is no RCD in this type of system."
However, the KT64 MFT uses a slighly different method for Line-Neutral impedance (from the KT64 manual):
"Measurement of Line – Neutral Loop Impedance and PSC:
Press the Mode Switch (F1) and select L-N/L-L or PSC. Connect the blue N lead of the ACC065 to the neutral of the distribution board, the brown L lead to one line of the distribution board. (See Fig.24)."
So that truely is "2 wire" rather than "3 wires going to 2 points" like on the old KT45.
So it might be that Kewtech have taken a retrograde step, and 2 point L-E loop test (when no N is available) is no longer possible.
But I suspect only they will be able to tell you (unless someone's got one and is willing to try).
Give em a buzz