I appreciate everybody’s reply’s and concerns. They’re all valid in their own right. Thank you.
New HSE regulations for x-ray machines in vetinary surgery’s.Out of interest, what specific regulation are you trying to comply with here?
Off the bat there would seem to be a few cheap solutions as oppose to building a monitoring circuit, an example been redundancy, why not just have 2 lamps and when one fails it is clearly obvious and maintenance will repair it or the user trained to report it.
That aside the current monitoring option seems the best as previously mentioned but my concern is your reliance on a contactor to isolate a power outlet, if this is for safety then by design it needs to be fail safe and have redundancy in the design thus the contactor needs monitoring as well via a safety control circuit... when you are trying to comply to machine code standards or similar standards then what often seems a simple solution can require a expensive and complex system to ensure safe use and operation.
It’s an old school surgery that just had a red lamp in a batten holder. The X-ray machine just on a 13amp standard plugtop. Warning notices present.Is this a first installtion of it's type?
I would have thought that hundreds of these machines would be in use and if it's goverened by legislation there would be an approved and tested method of providing the socket supply.
There must be a similar system in Hospitals with the Danger Xray in use sign above the door, presumably if that indicator fails then the equipment must be disabled.
As above somewhere a backup indicator would seem the simplest choice with a warning notice.
I know exactly what’s is happening, I’ve mentioned the very high resistance of the coil/ acting as a resistor. I’m looking for a soloution
Yep spot on apart from the contactor does pull in.You've got a 230v lamp in series with, I'd imagine, a 230v contactor coil. You will have approx 115v across each of them if you are powering them from a single phase supply. That's why your lamp is dim and contactor won't pull in.
In you want them to work in series you would need a lamp and contactor coil rated for around 120v each. If you can't find a contactor of that rating then you could use a 120vac relay to switch the 240vac contactor coil
Not quite or you wpYou've got a 230v lamp in series with, I'd imagine, a 230v contactor coil. You will have approx 115v across each of them if you are powering them from a single phase supply. That's why your lamp is dim and contactor won't pull in.
In you want them to work in series you would need a lamp and contactor coil rated for around 120v each. If you can't find a contactor of that rating then you could use a 120vac relay to switch the 240vac contactor coil
Yep spot on apart from the contactor does pull in.
Nope. Only if they have the same resistance, which they won't. You can't even guarantee that a 120V lamp and 120V contactor of equal power would work, because an AC contactor impedance and power factor vary with armature position, so the lamp would receive an impulse of more than 120V during pull-in, possibly blowing it, while its resistance would shoot up reducing the voltage available to pull the contactor in. These are not simple resistors that will divide the voltage ratiometrically.You will have approx 115v across each of them