Superb advice everyone. This particular forum is a great experience for a first time user. I found a German device that senses 0.02amps to 16amps. The 15w lamp will use 0.06amps @230volts so in theory it should work a treat.



There does not seem to be any CT chamber so I gather that terminals k and l would be wired in series with the load of the lamp. L\N line and neutral permanent supply. Any views on this device?

99A92C44-E1F4-4CE0-81BB-F0CE719C12E8.png
 
Last edited:
That would work.
No doubt I'm missing something here but the ones I've done, are switched from the surgeons panel and just come on at the same time as the socket, not on the control circuit. I don't think there is a requirement to provide a fail safe mechanism. A buzzer might be a better warning than a lamp.
 