The NIC are happy with us going for a longer period between factory calibrations if we routinely check the testers in our workshop more often and cross check them against each other. FWIW the type of work we do is quite easy on the MFTs, they don't get lugged about too much as once on a job they will tend to stay there for some time, they get shipped in flight cases and generally pampered. Also, most of our resistance and impedance figures are known in advance so we tend to spot a suspect reading when it happens. In other words the potential risk from calibrating less often is more than compensated by the way we use the testers. Any sign of trouble, if one instrument disagrees with another or turns in more than a couple of odd-looking results, then off it goes to the cal house.
I agree, decent instruments can be impressively stable. I have two identical Fluke multimeters, both 18 years old. One has been in my toolbox, used daily, been round hundreds of baggage conveyors in dozens of airports, submerged in water, buried in sand in the Sahara and in snow in Siberia. Unlike the MFTs It has seen a bit of the world. Its twin is the backup, lives in a cupboard, never done a day's work in its life. The two get compared to a precision meter on the bench every 6 months and neither has ever been adjusted - they still agree to better than 0.1% on DC volts.