basset
Whilst performing various continuity tests my Di-Log 9083P is only showing zero 0:00. any help would be most appreciated.
Sophie
Sophie
^^^ agreed... replace batteries, and fuse some times....!! un-null leads then start again!! good piece of kit just can get awkward at times..!
I've replaced the batteries as you've advised but still zero results. Any idea where would i get the fuse from?
thanks
Sophie
Hi Sophie, You require a T3.15A, 500V 1.25" HRC Ceramic Fuse such as SIBA List No. 70 065 65 similar to those also used in the Fluke 165xB series MFTs. When using the link below ensure you select the correct fuse from the drop down list before purchasing:
FLUKE 1651,1652,1653 & B TESTERS REPLACEMENT FUSES | eBay
Just ordered a couple
Thanks for the link
Hi Sophie did you manage to solve your problem? I'm having the same issue with my tester!!