[FONT=&]The Error 5 noise warning is displayed when there is excessive noise on the supply. Noise or transients on the supply voltage effect the accuracy of the measurement during the test and introduce an error in the reading. When excessive noise is detected by the Fluke 16xx series MFT it automatically inhibits the test and displays the Error 5 warning. The noise or transients on the supply voltage are usually caused by active appliances or equipment on the supply, the source is usually on the installation being tested, but equally may be caused by equipment on another installation connected to the same supply source.
Isolate (switch OFF) all appliances and equipment connected to the installation, particularly automatically switched loads (eg fridges/freezers/boilers) and motor controllers and retest.
However, sometimes even this will not solve the problem, because noise or transients external to the installation are the cause and it may require testing at a different time of day to avoid the active period, for example, during their lunch hour or after normal working hours eg. garage down the road maybe using welding equipment or the Vets /Doctors surgery using X-ray equipment.[/FONT]
In an industrial environment where continuous production must be maintained and the noise source cannot be isolated, often the only thing you can do in the short-term is to take a series of readings and record the average value to minimise the error. However, because the Fluke automatically inhibits tests in noisy conditions this may require the use of a different test instrument. Later retesting for accurate results during a planned maintenance shut-down period.