Here's a tip for anyone still using an old MFT1700 series tester. Mine dates from 2012.

I found about three years in - by 2015 in this case, that the zeroing of the leads on my MFT1720 was becoming increasingly random and the zero would wander about by a fraction of an ohm which became irritating and time consuming. Others talk of replacing leads -(Kewtech are good, and this is certainly worth doing, but it's's not the only problem). At the end of the day, a copper lead is just that. there's no magic. It's the contacts which make the difference.

Cleaning the external leads and instrument terminals with Isopropyl alcohol and cotton buds helps. - But the big win on these ageing instruments is the contact between the terminals "pod" and the main board.

So if you are interested, here's what you do.
Turn the instrument over, remove the batteries, contacts towards you. Remove the two screws adjacent to the fuses.
In between the two fuses you see a rectangular hole and a square one. The rounded plastic boss you can see down the rectangular opening is a robust plastic part of the banana-terminals"Pod". The Pod is the oval-shaped moulding which holds all contacts facing your tummy now. The rear of the terminals are socketed into the main board with male gold-plated spring pins.

Having removed those two screws, nothing except friction holds the "pod" into the case.

So, get a wide blade flat screwdriver and insert it into the rectangular hole, then prise it to push the grey "boss" (part of the oval contacts pod) out of the front of the instrument. You may find first time it's really tight, and one end of the pod wants to release firsts. Hold it straight and lever it out. First time will possibly need quite a bit of force.

The moulding comes right out, and bingo, the source of your problem. You will see spring gold plated pins, which in the factory are lubricated with some Contra-Lube (or similar grease).
These gold contacts can get a little grubby, and a small build-up will add resistance and cause your never ending zeroing problems.
Clean them a little very gently, - if you like with a little isopropyl alcohol. Add a dab of Contra-Lube and push the pod back in.

Bingo, prefect zeroing. I find annual of six-monthy maintenance helps, but I will always do it if I feel the zeroing is beginning to wander a little.
 
still does not get round the problem of being unable to clean the ends of the leads as they are now secreted within shrouds and inaccessible.
 

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Green 2 Go Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses Heating 2 Go Electrician Workwear Supplier
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Advert

Daily, weekly or monthly email

Thread starter

1Justin

Arms
-
Joined
Location
Surrey
If you're a qualified, trainee, or retired electrician - Which country is it that your work will be / is / was aimed at?
United Kingdom
What type of forum member are you?
Practising Electrician (Qualified - Domestic or Commercial etc)
Business Name
Circitas Ltd

Thread Information

Title
MFT1700 series wandering zero problems fixed.
Prefix
N/A
Forum
Electrical Tools and Products
Start date
Last reply date
Replies
3

Advert

Thread statistics

Created
1Justin,
Last reply from
1Justin,
Replies
3
Views
1,219

Advert

Back
Top