Hi all,

I've always understood that using the anti-trip setting will normally give an inaccurate reading. However I wasn't so aware of level of it's inaccuracy..
My own tester is a KT64.When I've done a comparison test on it the difference between the non-trip and normal reading many times, and the difference in the reading was only ever up to a maximum of 0.1 ohm.
Today I finally received my Megger from one of my employers, and so I had a play with it today and thought I would check for the difference in readings on this kit.

I measured a Ze of 0.16 ohms, and when I turned it to anti-trip it jumped up to 0.63 ohms. I was somewhat amused that this kit could really apply nearly half an ohm in inaccuracy?? I re-tested a couple of times (allowing a bit of time between tests) and it maintained these readings.

I'm not necessarily concened about this, but I thought I would post on here to see if anyone else has checked out their tester and if the readings are similar to this??

:confused:
 
Had the same prob with 2 of our megger 1553's. (new at time 2 years ago)
Done the same as yourself and tested a non RCD protected circuit on the high setting and then again on the no trip both meters showed the same inaccuracies of about 0.5 ohms.
Both were sent back to megger and repaired results now are within 0.02 ohms and remain stable as i check them monthly as routine.
 
If I get anyone else with a similar opinion I'll be sending it back.



I wonder if the is a written maximum allowance for this kind of inaccuracy somewhere.
 
Last edited:
I used to have some Megger seperates.

Did exactly the same thing.

Im sure somewhere in the manual it does state this.
 
I used to have a megger CM500, did exactly the same thing. I spoke to megger and they said it was within there tolerances. I changed to a 1552 and since then everything is fine.

Ian
 