• Please checkout this (ad-free) update post I've done on a spare blog. Feel free to bookmark it and check everynow and again for update if ads are pissing you off. Always feel free to ad-block them for now though if you want. I'll let you know when you can disable the ad blocker again, only if you want to see the ads. They should look nice and neat in the end.

Discuss PAT Testing fail query in the Electrical Testing & PAT Testing Forum area at ElectriciansForums.net

Reaction score
2
Welcome to ElectriciansForums.net - The American Electrical Advice Forum
Head straight to the main forums to chat by click here:   American Electrical Advice Forum

Hi,

I perform annual PAT checks for my company (office and workshop). The Apollo 400 machine we use has recently gone for calibration and service and since it's come back, items which passed easily last year (Earth Continuity below 0.1ohm), are now not passing. I've now raised the earth continuity limit to factor in cable resistance etc. but why is it that some items which were say 0.8ohm are now 0.17ohm and wont pass? Any ideas?

Many thanks,
 

littlespark

-
Esteemed
Arms
Reaction score
16,340
Because it wasn’t calibrated.

Out of calibration means that something is off inside the machine which can give false readings.
When it goes to be calibrated, these “off” things are adjusted against test examples with known good values…. Which brings the results back to being more accurate.
 
OP
R
Reaction score
2
Because it wasn’t calibrated.

Out of calibration means that something is off inside the machine which can give false readings.
When it goes to be calibrated, these “off” things are adjusted against test examples with known good values…. Which brings the results back to being more accurate.
Hi. We have annual calibration certs since we purchased the machine.
 

DPG

-
Esteemed
Arms
Patron
Reaction score
16,305
Ah fair enough. Have you got a decent multimeter you could use to verify the results. Vital to get good connections obviously when you're measuring such low resistances.
 

Avo Mk8

-
Esteemed
Reaction score
1,502
If you keep your PAT results from previous years, can you review these for evidence there has been a step change in the Apollo results after calibration, if that is what's happened.
e.g. If you have results for workshop equipment that showed a value of around 0.08 ohms year on year for 3 or 4 years, and now the value is 0.17 ohms, and you then look at the results for more equipment, and they have all done the same thing (even better if they are now all around 0.09 ohms higher), then you have solid evidence your Apollo is suddenly producing consistently different results from all previous years, and I would send it back for investigation.

It's not that unusual for equipment to get upset in transit, or less likely for an error to be made during calibration.
If the data gives evidence there's a consistent change in the Apollo's results, the calibration company ought to look into it for you.
 
OP
R
Reaction score
2
If you keep your PAT results from previous years, can you review these for evidence there has been a step change in the Apollo results after calibration, if that is what's happened.
e.g. If you have results for workshop equipment that showed a value of around 0.08 ohms year on year for 3 or 4 years, and now the value is 0.17 ohms, and you then look at the results for more equipment, and they have all done the same thing (even better if they are now all around 0.09 ohms higher), then you have solid evidence your Apollo is suddenly producing consistently different results from all previous years, and I would send it back for investigation.

It's not that unusual for equipment to get upset in transit, or less likely for an error to be made during calibration.
If the data gives evidence there's a consistent change in the Apollo's results, the calibration company ought to look into it for you.
Hi Avo,

Thanks for the reply. Think that's what i'll end up doing. The cal cert says everything is fine but i'll get a better picture once start comparing back.
 

Reply to PAT Testing fail query in the Electrical Testing & PAT Testing Forum area at ElectriciansForums.net

Similar Threads

Hi All I’ve come across this heater Ariston Andris Lux Undersink Water Heater 2kW 10Ltr - Screwfix -...
Replies
14
Views
2K
Hello, I am currently learning and have come across PAT testing an old chandelier. On modern lamps I would use the plug and probe on the...
Replies
1
Views
1K
Hi I wondered if I could ask for advice on how to succeed in getting some basic technical information from the manufacturer of consumer...
Replies
27
Views
2K
Just started PAT testing and trying to understand why stuff fails. Recently had a large fridge fail. The cable was scuffed with a slice of the...
Replies
7
Views
2K
NAPIT Certification Scheme Pat Tester and testing help please
Hi all, new to this forum and could do with some advice re Pat Testing. As the new law relating to EICR's to the private rented sector is coming...
Replies
4
Views
3K

New Posts (Please Reply)

This website was designed, optimised and is hosted by Untold Media. Operating under the name Untold Media since 2001.
Top