Hi

Happy new year to everyone.

I have mft1720 with megger fused leads protected by 50mA fused.

On Zdb (not main but garage unit) I wanted to do a Z measurement, set it to high current mode (L-L/L-N and Z) when I attempted to take a reading between live wires it blew the fuse in the Netural wire? I believe you must use high current mode for doing Ze or Zdb, so why should the fuse blow?

cheers

krish
 
not sure exactly how you were doing the test, but sounds like you got something wrong.
 
GS38 compliant probes use 500mA fuses and have an appreciable resistance making them unsuitable when loop testing on live systems.

The fused lead itself will add a small variability into the test result.

Test leads fused at 10A will provide protection from short circuit faults but will not fail under normal testing conditions.

500mA Fused test leads (GS 38) order code 6220-789

10 A Fused leads order code 6220-827
 
Trypod, thanks man for this, I thought it might be something like this, so I will have to order the 10A fused leads, just wandoering if I could cheat and stick a 10A fuse in the GS38 leads I used?

krish




GS38 compliant probes use 500mA fuses and have an appreciable resistance making them unsuitable when loop testing on live systems.

The fused lead itself will add a small variability into the test result.

Test leads fused at 10A will provide protection from short circuit faults but will not fail under normal testing conditions.

500mA Fused test leads (GS 38) order code 6220-789

10 A Fused leads order code 6220-827
 

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Green 2 Go Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses Heating 2 Go
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Advert

Daily, weekly or monthly email

Thread Information

Title
Megger MFT1720 loop impedenace test issue
Prefix
N/A
Forum
Periodic Inspection Reporting & Certification
Start date
Last reply date
Replies
3

Advert

Thread statistics

Created
krishcanag,
Last reply from
krishcanag,
Replies
3
Views
4,710

Advert

Back
Top