Hi All,

Anybody have a METREL meter thats trips RCD on triplock setting when Zs is taken?

The fault seems exclusive to the probes lead. The coiled socket lead is not affected and is producing normal readings.

My friend also has an earlier METREL meter (not sure what model) which has exactly the same problem.

His meter passed calibration which l dont understand how, as he has this fault that occurs constantly.

Mine has only just started doing so and ive checked all the settings with 30mA definitely selected on the triplock setting.

Advice anyone plz? May be tomorrow when l check feedback so ill say thanks right now and will get back to anyone that replies asap.

Thanks Jabbajaws.. .:cry_smile:
 
My Metrel tested the Zs (on trip lock) once and it was OK, I then did some work and switched everything on and did the same test which then tripped the RCD, so I assumed this was from additional leakage currents from other circuits pushing it over the trip limit (this was an up front 30mA RCD).
 
Thanks for the usual swift replies. Although l normally short neutral and earth together, as instructed to do by college tutors, 4 years ago, l have attained satisfactory readings using this method on numerous occasions. And besides how often are is it even possible to perform a 3 lead test on for instance a light switch? There's normally no neutral (unless it has been carried through the switch). I have never actually performed a 3 lead Zs test using the probe leads on this meter, as l have never had a reason to do so, but l will attempt this out of curiosity, at a location where possible and report back my findings. Thanks... JJ
 
Thanks for the usual swift replies. Although l normally short neutral and earth together, as instructed to do by college tutors, 4 years ago, l have attained satisfactory readings using this method on numerous occasions. And besides how often are is it even possible to perform a 3 lead test on for instance a light switch? There's normally no neutral (unless it has been carried through the switch). I have never actually performed a 3 lead Zs test using the probe leads on this meter, as l have never had a reason to do so, but l will attempt this out of curiosity, at a location where possible and report back my findings. Thanks... JJ

You can do the ZS with earth and neutral linked together on a circuit without an RCD but with an RCD you will have to use the three leads so I'm afraid it get your steps off the van, take the light fitting cover off and then try and hold all three leads, the tester, press the test button and still maintain 3 points of contact on your steps and hold your cup of tea at the same time!!! O and remember it takes the tester about 3 Hours !! to perform the trip lock test.
Or you could just do A R1 + R2 test it at the light switch and Calculate the ZS
 
You can do the ZS with earth and neutral linked together on a circuit without an RCD but with an RCD you will have to use the three leads so I'm afraid it get your steps off the van, take the light fitting cover off and then try and hold all three leads, the tester, press the test button and still maintain 3 points of contact on your steps and hold your cup of tea at the same time!!! O and remember it takes the tester about 3 Hours !! to perform the trip lock test.
Or you could just do A R1 + R2 test it at the light switch and Calculate the ZS
Sounds like l need a wander lead to perform this action. Thanks for the useful advice.. Sorry for cramped text. My enter key has stopped working, lol...
 
Sounds like l need a wander lead to perform this action. Thanks for the useful advice.. Sorry for cramped text. My enter key has stopped working, lol...

You don't need a wounder lead , turn off circuit and and lock it off check circuit is dead at light switch connect the earth to the live and do the R1+R2 test at the CU and then calculate ZS
 
Last edited:
I have this meter and i always use three fingers of one hand to hold probes and other hand to press test. That's where megger are good with the auto-start thing.
 

Reply to METREL METER FAULT? METREL Easitest MI3000 in the Electrical Tools and Products area at ElectriciansForums.net

Electricians Tools | Electrical Tools and Products

Thanks for visiting ElectriciansForums.net, we hope you find the Electricians Tools you're looking for. It's free to sign up to and post a question yourself to find a tool or tool supplier either local to you, or online. Our community of electricians and electrical engineers will do their best to find the best tool supplier for you.

We also have a Tiling Tools advice from the worlds largest Tiling community. And then the Plumbers Forums with Plumbers Tools Advice.

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc
This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock