I asked a guy that works with me to get his tester out aswell, which was only a week old. His tester was working accurately.
We done a loop test using the megger and his di-log on either side of the rcd. The megger recorded .50ohm difference, where as his tester recorded no difference at all.
It’s on both tn-s and tn-c-s. I’m yet to use it on a tt system.
I have spoken to megger technical. They said to use the hi current reading and just to do a ramp test to ensure the rcds are tripping at an acceptable Ma. Just seems bizarre it’s with both my 1552 testers
the Rcds are MK mainly but it’s recently happend on a CPN rcd aswell. It seems to be more on the recent metal clad dbs, I don’t know whether they have altered the rcds in anyway
Hi guys, I own 2 megger mft 1552 testers.
I recently found that loop readings on both testers give me high Zs readings on final circuits protected by 30ma rcds on low current.
If I select the hi current setting I get a reading I expect to achieve, but the hi current test never trips the rcd...