Hello

I've got a Fluke 1621652B which I've been using without fault for the last couple of years.


Was doing some fault finding on a job this morning, random RCD tripping issue, when I was using the ramp test function to check the installed RCD.

Tried first test with all load disconnected to test RCD and result was fine. Reconnected all outgoing circuits and turned basics on (fridge, etc...), tried to ramp test from a local socket, RCD tripped instantly, Fluke had nothing on the screen (no result or anything).

I then tried to use the fluke again to which it came up with 'FUSE' on the screen. I thought this was strange however attempted to replace the internal fuse.

I didn't have any of the F1A fuses handy however had an F10A fuse of same size which wasn't blown (verified with voltstick continuity test) so tried this in order to verify there was no problem with tester (original fuse had in fact blown)

Unfortunately the tester still comes up with the FUSE fault, even with a working fuse fitted. There is a loud click from within the tester before the FUSE fault comes up.

Whats the prognosis? Is it knackered? Does anyone have any ideas of costings to repair? I've got solid EICRs booked in for the next few days and could really do without this hassle, should have had a spare!
 
Hello

I've got a Fluke 1621652B which I've been using without fault for the last couple of years.


Was doing some fault finding on a job this morning, random RCD tripping issue, when I was using the ramp test function to check the installed RCD.

Tried first test with all load disconnected to test RCD and result was fine. Reconnected all outgoing circuits and turned basics on (fridge, etc...), tried to ramp test from a local socket, RCD tripped instantly, Fluke had nothing on the screen (no result or anything).

I then tried to use the fluke again to which it came up with 'FUSE' on the screen. I thought this was strange however attempted to replace the internal fuse.

I didn't have any of the F1A fuses handy however had an F10A fuse of same size which wasn't blown (verified with voltstick continuity test) so tried this in order to verify there was no problem with tester (original fuse had in fact blown)

Unfortunately the tester still comes up with the FUSE fault, even with a working fuse fitted. There is a loud click from within the tester before the FUSE fault comes up.

Whats the prognosis? Is it knackered? Does anyone have any ideas of costings to repair? I've got solid EICRs booked in for the next few days and could really do without this hassle, should have had a spare!


Contact FLUKE and see what they say, better still take it to your local wholesaler and get them to send it off for repairs
 
I actually did call FLUKE who's engineer was supposed to be calling back but the call never came

Even better, took it to a local company who actually do the repairs themselves rather than a wholesaler, it turned out to be loose contact springs on the fuse, apparently if there is a high resistance joint the tester sees it as faulty

The guy even threw in a few spare F1A fuses

If anyone is from the Yorkshire area, PJO industrial in Barnsley are a good company to do business with, saved my skin today!
 
Last edited:
  • Like
Reactions: 3 people

Similar threads

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electric Underfloor Heating Electrician Courses Heating 2 Go Electrician Workwear Supplier
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Advert

Daily, weekly or monthly email

Thread starter

Joined
Location
West Yorkshire

Thread Information

Title
Fluke 1652B "FUSE" fault? Urgent!
Prefix
N/A
Forum
Periodic Inspection Reporting & Certification
Start date
Last reply date
Replies
2

Thread Tags

Advert

Thread statistics

Created
ray-bentos,
Last reply from
ray-bentos,
Replies
2
Views
9,645

Advert