Could someone tell me if it is possible in fluke 1663/1664 to activate the option to carry out an insulation test without having the finger fixed on the test button? I am not referring to the differential test, which I know can be done by detecting voltage and not pressing the "start". I only mean insulation test... or maybe it's not possible...? (it is not in the manual)
The only conclusion I can draw is for the safety of the operator... Other manufacturers, which are many, do the test by pressing the insulation test once...
 
I have testers that take both approaches. Metrel is press and let go and it takes samples over a couple of seconds. Meggers are press and hold.

The advantage of those that keep the current running while you hold is that you can see if the reading is slowly improving or vice versa. Sometimes you can see lower IR which improves as you are physically drying out the cable by running current through it.
If anything, I consider the method your tester uses to be the superior method!
 
Last edited:
I have testers that take both approaches. Metrel is press and let go and it takes samples over a couple of seconds. Meggers are press and hold.

The advantage of those that keep the current running while you hold is that you can see if the reading is slowly improving or vice versa. Sometimes you can see lower IR which improves as you are physically drying out the cable by running current through it. I
If anything, I consider the method your tester uses to be the superior method!
Cleared up, thanks a lot Timhoward.
 
Meggers also have the facility to lock the test button in the hold position. I don't know if the Flukes do.
 
Meggers also have the facility to lock the test button in the hold position. I don't know if the Flukes do.
It does not appear in the manual, I even called the technical service to confirm it... It does not have it.
 

Similar threads

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electric Underfloor Heating Electrician Courses Heating 2 Go Electrician Workwear Supplier
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Advert

Daily, weekly or monthly email

Thread starter

Joined
Location
SPAIN
If you're a qualified, trainee, or retired electrician - Which country is it that your work will be / is / was aimed at?
All Other Countries (This Is English Speaking Website Only - WE don't mind Google Translate Users :)
What type of forum member are you?
Other

Thread Information

Title
Insulation fluke 1663/1664
Prefix
N/A
Forum
Electrical Tools and Products
Start date
Last reply date
Replies
4

Advert

Thread statistics

Created
sldz1200,
Last reply from
sldz1200,
Replies
4
Views
1,239

Advert