Folks.
I need to install a defib unit outside on a wall.
Does this require an RCBO or not, just wondering as the unit has a small heater within and should it trip out for any reason other than under fault conditions (it is in a dampish environment, a harbour) it will get damp.
So MCB or RCBO.
Many thanks
AJ.
 
What do the manufactures instructions say?
 
All the same regulations apply as they do to any other circuit.

Go through the list of reasons why RCD protection may be required and if none of them apply then it is not required.
 
The link mentions RCD protection at note 10...

But the 4A (time-lag) fuse spur seems a bit strange. Why not a regular 3A?
 
The link mentions RCD protection at note 10...

But the 4A (time-lag) fuse spur seems a bit strange. Why not a regular 3A?
Probably because there is a possibility of a short overload.
 
  • Like
Reactions: nicebutdim
Just a thought and probably not relevant but there are some exceptions in the regs regarding medical equipment,
cant remember exactly what they are but on occasions protection may be omitted or reduced to ensure the device remains functional.
or am I thinking of lifting magnets? its been a long day but there are a few things to look up in the bible if any one is interested.
 
The power required is not for medical equipment… it’s for a light and a heater.

The defib itself is self contained battery powered.
 
  • Like
Reactions: nicebutdim
agreed, however, it may be considered an essential supply as without it the medical equipment may not work?
playing devils advocate here!
 
Me thinks this is being over thought. As @davesparks suggests in #5 this is the correct method to follow. Something which may get damp needs to be installed to prevent this and not compensate by fitting an RCD which is not a solution.
 
I'm aware of 3 defibs local to me that have been vandalised over the years, doors kicked off etc, leaving the heaters and guts exposed, so whenever i do them i have an rcd.
 
  • Like
Reactions: alexander jackson
Most instructions, including the one linked to earlier, ask for rcd protection… so that’s the answer in this particular instance.

Whether regs demand it or not, we must take manufacturers instructions into considoration.
 

Similar threads

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electric Underfloor Heating Electrician Courses Heating 2 Go Electrician Workwear Supplier
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Advert

YOUR Unread Posts

Daily, weekly or monthly email

Thread starter

Joined
Location
uk

Thread Information

Title
RCBO or not?
Prefix
N/A
Forum
UK Electrical Forum
Start date
Last reply date
Replies
15

Advert

Thread statistics

Created
alexander jackson,
Last reply from
littlespark,
Replies
15
Views
1,187

Advert