I suspect my Megger 1710 tester is faulty
So I cut exactly 1 meter of 2.5 -1.5

I have nulled the probes as per the instructions.

Continuity test.

End to end
L = 0.11 Ohms
N = 0.11 "
CPC - 0.11 "

Question is this test valid.
Why is the CPC not 1.67 times 0.18 Ohms

What could I do to prove it's faulty or not before sending it off.

ADDED new duracells put in

Thanks in advance of any help.
 
Last edited by a moderator:
all those readings are far too high for 1m . try a different set of leads for starters and make sure the plugs/sockets on leads and meter are clean. another tip, when reading such low values, don't null the leads. subtract from the result after.
 
Last edited:
  • Like
Reactions: 1 person
Thanks for the tips - the meter is virtually brand new & clean I got it 11months ago used it 3 times.
Bought the kewtech light testing KIT and that didn't work, it doesn't work with the megger supplied plug tester
or the kewtech one, I put that down to incompatiblity.

Gonna send it back before the warrantly runs out. I'm guessing I'll ring megger on monday to find their off till after xmas :sad_smile:

I work as an electricains labourer and only 4 weeks of that was domestic in the past year.
No testing whatsover.
Need to get it sorted as I have guranteed min 3 weeks domestic with a local firm.
Only got this cos I was the one who got the job.
 
my lightmates don't work with megger test leads either. you need to get it off to megger asap, get it fixed under warranty.
 
i had a 1552 that was the same. all other functions were spot on, but continuity was hit and miss, mostly miss.
 
Going back then, I will do the RMA today hopefully that should keep me within the warranty time.
 
what is RMA?
 
Posted to Megger, it was signed for on 24/12/15 @ 9.02am they say 5 working days for return. :D
Lucky I noticed it cos it was bought on 12 jan 2015 and I didn't register for the 3yr warranty.
 
Yes! Got it back today the service report says:
Faulty current circuit.

They've put a date sticker on it saying it's been calibrated....result.
Bit miffed I have a repaired tester, would I have a resonable shout on insisted in a new replacement ?
 
It has to be calibrated after being repaired, so I'd be worried if they hadn't done it!
 
yes! Got it back today the service report says:
Faulty current circuit.

They've put a date sticker on it saying it's been calibrated....result.
Bit miffed i have a repaired tester, would i have a resonable shout on insisted in a new replacement ?
the termsof your guarantee would be standard.manufacturer repairs or replaces athis discretion.
 
the termsof your guarantee would be standard.manufacturer repairs or replaces athis discretion.
I might float the ideal past them that I think I should have a new machine, see how they react.
It's had less than 2hours 'on' time, God's honest. :-)
 

Similar threads

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electric Underfloor Heating Electrician Courses Heating 2 Go Electrician Workwear Supplier
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Advert

Daily, weekly or monthly email

Thread Information

Title
megger 1710 problems or not
Prefix
N/A
Forum
Electrical Course Trainees Only
Start date
Last reply date
Replies
13

Advert

Thread statistics

Created
newtothis,
Last reply from
newtothis,
Replies
13
Views
133

Advert